Author Topic: Rigol DS1054Z -- Features, Usage, and... BUGS !!  (Read 23519 times)

Hero Member
*****

Posts: 922
Yes, it sounds to me like the scope is just waiting for a triggering event.  Brad, can you post a screenshot or a short video of the scope when this happens?

In the Trigger section on the front panel, the "Mode" button switches between the three Sweep Modes: Auto, Normal, and Single. When the Mode is Auto, the scope free-runs and shows a trace even if it is not triggered. When the mode is Normal, the scope waits for a trigger event before it will update the trace on screen. When the mode is Single, it waits for a trigger event then does one full sweep and Stops.

If the scope is in Auto sweep mode and you are getting that "Wait", then there might be something wrong. But if it is in Normal sweep mode, it is just telling you that your Trigger parameters aren't satisfied.

(This is different from the "Auto" setup button in the very top row of buttons! I am talking about the Auto Sweep Mode selected in the Trigger section of the panel, not what happens when you push that "Auto" setup button!)

By "Trigger Event" we mean something that satisfies what you have set in your Trigger menu and the vertical position of the Trigger Level indicator. So for example say you are looking at a Sine Wave of varying amplitude, and you are in Normal sweep mode, and your Trigger Type is set to "rising edge" and your Trigger Level is set to + 1 volt. The scope will Wait until the signal amplitude is over 1 Volt and will only trigger when it is over. When the signal level drops below 1 volt the scope will stop sweeping and will be "Wait"ing for the next time the signal exceeds the 1 volt setting. And it will continue to show a "frozen picture" of the last triggered sweep, unless you push the "Clear" button in the top row.

This should not be any different than how the old Atten or any other DSO behaves though.


Brad we should try to get this issue sorted right away, so that IF there actually IS something wrong with the scope, you can return it under the Vendor's warranty and get a replacement immediately.




---------------------------
"The easiest person to fool is yourself" -- Richard Feynman
   

Hero Member
*****

Posts: 1236
Well,not sure what you mean !waiting! for the triggering event.
It means that it is waiting for the condition defined in the trigger menu, to be satisfied.
« Last Edit: 2017-05-19, 21:38:32 by verpies »
   

Hero Member
*****

Posts: 922


---------------------------
"The easiest person to fool is yourself" -- Richard Feynman
   

Sr. Member
****

Posts: 448
Maybe this will help:

http://www.youtube.com/watch?v=ieCQGFWAitk
Ahhh I'm yet to get out of auto mode. That helped me understand lots of things . Thanks.
   

Sr. Member
****

Posts: 448
Ok I don't get this . My TBP are stacked with a ferrite core.
   
Group: Elite
Hero Member
******

Posts: 3052
It's turtles all the way down
Ok I don't get this . My TBP are stacked with a ferrite core.

Dear Jim

It looks correct, you are reading input current pulses across an inductance (primary TBF, yellow ch), not seeing peaks because pulses somewhat smoothed by the two capacitors, C1, C2.

The current pulses are coupled to the unloaded secondary of TBF, blue trace.

Perhaps someone else could also describe. What is the question?

Regards


---------------------------
Just because it has a patent application or is patented does not always mean it really works.
   

Hero Member
*****

Posts: 922
Sharks and elephants.

Your sharks are eating your elephants.   


---------------------------
"The easiest person to fool is yourself" -- Richard Feynman
   

Sr. Member
****

Posts: 448
Mm I though the elephants were trampling the sharks. I thought the yellow should leading and it looks like it's trailing. Thought I may have a setting wrong.
   
Hero Member
*****

Posts: 1560


Buy me some coffee
Yes, it sounds to me like the scope is just waiting for a triggering event.  Brad, can you post a screenshot or a short video of the scope when this happens?

In the Trigger section on the front panel, the "Mode" button switches between the three Sweep Modes: Auto, Normal, and Single. When the Mode is Auto, the scope free-runs and shows a trace even if it is not triggered. When the mode is Normal, the scope waits for a trigger event before it will update the trace on screen. When the mode is Single, it waits for a trigger event then does one full sweep and Stops.

If the scope is in Auto sweep mode and you are getting that "Wait", then there might be something wrong. But if it is in Normal sweep mode, it is just telling you that your Trigger parameters aren't satisfied.

(This is different from the "Auto" setup button in the very top row of buttons! I am talking about the Auto Sweep Mode selected in the Trigger section of the panel, not what happens when you push that "Auto" setup button!)

By "Trigger Event" we mean something that satisfies what you have set in your Trigger menu and the vertical position of the Trigger Level indicator. So for example say you are looking at a Sine Wave of varying amplitude, and you are in Normal sweep mode, and your Trigger Type is set to "rising edge" and your Trigger Level is set to + 1 volt. The scope will Wait until the signal amplitude is over 1 Volt and will only trigger when it is over. When the signal level drops below 1 volt the scope will stop sweeping and will be "Wait"ing for the next time the signal exceeds the 1 volt setting. And it will continue to show a "frozen picture" of the last triggered sweep, unless you push the "Clear" button in the top row.

This should not be any different than how the old Atten or any other DSO behaves though.


Brad we should try to get this issue sorted right away, so that IF there actually IS something wrong with the scope, you can return it under the Vendor's warranty and get a replacement immediately.

All good,was user error lol.

Now set to auto,in stead of normal O0


Brad
   

Sr. Member
****

Posts: 448
I don't understand what I'm looking at here obviously it's not 412v so what is it actually measuring? I have the yellow channel across the pulse TBP coil and the blue across  two TBP coils wound in series either side of the pulse coil with a ferrite rod through the centre. So a transformer basically.
   
Hero Member
*****

Posts: 1560


Buy me some coffee
I don't understand what I'm looking at here obviously it's not 412v so what is it actually measuring? I have the yellow channel across the pulse TBP coil and the blue across  two TBP coils wound in series either side of the pulse coil with a ferrite rod through the centre. So a transformer basically.

I would say you have your scopes probe voltage set to 10x or 100x,instead of 1x  :D
   

Hero Member
*****

Posts: 922
It's not unusual to get really high voltages induced in the secondary, by fast rise/fall times in the primary.

But just to be sure, check that your switch on the CH2 probe itself is set to 10x and the CH2 channel setting is also set to 10x. The thickness of the trace in the third scopeshot is a little suspicious.


---------------------------
"The easiest person to fool is yourself" -- Richard Feynman
   

Sr. Member
****

Posts: 448
Yep I checked the probe switches before posting. Both 10X . Just wanted to check I wasn't missing something basic as is often the case :)
   

Hero Member
*****

Posts: 922
So, you have the channel set to 100 v/div and the automatic Measurement of Vp-p is showing the correct value of what the trace is showing.

I'm a little worried about the thickness of the trace, that's kind of unusual, but the high p-p voltage value in itself doesn't strike me as being strange at all.


---------------------------
"The easiest person to fool is yourself" -- Richard Feynman
   

Full Member
***

Posts: 223
Believing something false does not make it true.
So, you have the channel set to 100 v/div and the automatic Measurement of Vp-p is showing the correct value of what the trace is showing.

I'm a little worried about the thickness of the trace, that's kind of unusual, but the high p-p voltage value in itself doesn't strike me as being strange at all.

I guessing that if Jim shortens the time base to expand the wide trace horizontally he will see there is some RF or at least some high frequency component  riding on the main signal causing the wide scope trace.  I have seen that before.



---------------------------
Just because it is on YouTube does not make it real.
   

Sr. Member
****

Posts: 448
Thanks blokes! I will check after dinner :) Just cooking a roast with home grown veggies :)
   